Hey Randy,
I've read that you have one of the American Marine Nitrate Monitors but cannot comment on the accuracy. I'm intersted in purchasing one but would like to test the accuracy, any ideas on how I could go about doing this?
Even if it's not accurate but would at least pick up on increasea/decreases in a tank over time that would be benificial.
I"m thinking I could use the calibration solutions that come with it and try those on a few of the hobby test kits.
Randy Holmes-Farley
01/26/2007, 02:42 PM
It clearly responds to nitrate levels, but it is not the type of thing that you can leave in the water like a pH meter. It is more like a kit in that you take a water sample and test it.
The calibration kits that come with it, if made properly, are a good way to test test kits, and I've recommended that to others.
To test the meter accuracy, you'd have to make standards independent of the Pinpoint standards, so will need access to some good quality nitrate compounds,like sodium nitrate.
jamal-188
01/29/2007, 01:44 PM
You suggested before that you should probably calibrate these units at least once a day. Do you think it's necessary to calibrate if going from one tank to another?
Randy Holmes-Farley
01/29/2007, 02:24 PM
Not if you do them all in a span of a few minutes. But hours, yes, I'd recalibrate.
TandN
01/29/2007, 04:43 PM
Hey Randy just wondering your take on the monitor since you have one is it giving you a true reading or is it like test kits which give false readings?
loudell
01/29/2007, 06:29 PM
Hi Guys. Let me jump in with a few comments about the nitrate monitor and the calibration fluids. The meter will display an accurate and reproducible value for NO3 in seawater. It will tell you if your nitrates are 4 or 6 ppm, or 67 or 68 ppm NO3, etc. and if you want to verify of the meter is reading correctly you can put the probe back into the appropriate nitrate calibration fluid to verify calibration (just like a pH meter). The nitrate calibtation fluids are in a seawater matrix and it is not easy to make your own. Definately not Deionized water and sodium nitrate. The meter comes with 3 different seawater matrix nitrate fluids: 1 ppm, 10 ppm and 100 ppm. If your objective was to determine how accurate the nitrate monitor (or your colormetric test kit) actually is then you should be testing the nitrate solutions. The accuracy of the 100 ppm fluid is +/- 1 ppm NO3 while the accuracy of the 10 ppm fluid is +/- .5 ppm NO3, the accuracy of the 1 ppm fluid is +/- 0.1 ppm NO3.
Hope this clears up a few things...Lou Dell/American Marine Inc.
TandN
01/29/2007, 06:57 PM
Yes it does thank you Let me ask you are the calibration fliuds like the PH where you use it once and then throw it out ?
loudell
01/29/2007, 07:01 PM
The nitrate calibration fluids are not subject to the oxidation that pH fluids suffer. Nitrate Calibration fluids are very robust and are limited by technique contamination which basicly means that are as accurate as you are careful not to mix other fluids with them. The fluids that come with the meter are expected to last the hobbyist 6-12 months.
TandN
01/29/2007, 07:17 PM
Kewl thank you for the awnsers helps out alot :)
Randy Holmes-Farley
01/30/2007, 06:06 AM
Hey Randy just wondering your take on the monitor since you have one is it giving you a true reading or is it like test kits which give false readings?
IMO, if the standard is made out of water close to the tank water, it is likely to be reasonably accurate. One concern with ion selective electrodes in general is what other ions may interfere with it. I do not know how close these solutions are to typical tank water, a d what variables really matter, but it has the potential to be very accurate.
loudell
01/30/2007, 07:13 AM
Randy, The nitrate standards are painstakingly close to seawater and they were put together pretty much an element at a time. Some people may think it would be easy to mix deionized water, salt mix and sodium nitrate to create standards. Definately not true, trace nitrate can be found in salt mix and water supply as well. If you have the ability to accuraly measure the nitrate in the standards by another accurate method; please do so. The meter will be the most accurate method to precisely determine nitrate in a marine enviornment.
Randy Holmes-Farley
01/30/2007, 08:58 AM
Thanks, Lou. :)
Do you know off hand how far folks would need to deviate from NSW in their tanks before any ions became a problem? Does the probe manufacturer describe which ions are most interfering? They probably list a bunch of things we won't have, but might list some we could. My old Orion catalog (1992) lists perchlorate, iodide, chlorate, cyanide, bromide, nitrite, HS-, carbonate, bicarbonate, chloride, phosphate, acetate, fluoride, and sulfate. The question is how interfering they (or others) might be to your probe at levels likely to vary from NSW.
What about salinity of 30 ppt?
loudell
01/30/2007, 09:10 AM
Randy, the most interfering ion is definately sodium. The combination of the meters algorithm and calibration fluids counteract the impact down to about 21-22ppm with +/- about 2 ppm accuracy at the far end of that spectrum. Many people want to use the meter with fresh water and it will not be accurate there even if you formulate fresh water standards because the mathematical formula is optimized for marine saltwater.
Randy Holmes-Farley
01/30/2007, 09:54 AM
combination of the meters algorithm and calibration fluids counteract the impact down to about 21-22ppm with +/- about 2 ppm accuracy at the far end of that spectrum.
Do you mean 21-22 ppt?
loudell
01/30/2007, 10:00 AM
Sorry, you're right. I meant ppt.
Regarding electrode interference if you look at potential interference for pH electrodes the list is also long. But essentially if you look at the conditions that each ion will interfere and the concentration range and you carefully create calibration medium that takes these factors into proper account the result is a highly accurate measurement.
Expample: pH is highly temperature dependant however in the range of an average aquarium between 68 and 86 F the resulting interference is negligable.
mwitten
01/30/2007, 10:28 AM
<a href=showthread.php?s=&postid=9112477#post9112477 target=_blank>Originally posted</a> by loudell
Randy, The nitrate standards are painstakingly close to seawater and they were put together pretty much an element at a time. Some people may think it would be easy to mix deionized water, salt mix and sodium nitrate to create standards. Definately not true, trace nitrate can be found in salt mix and water supply as well. If you have the ability to accuraly measure the nitrate in the standards by another accurate method; please do so. The meter will be the most accurate method to precisely determine nitrate in a marine enviornment.
Loudell,
Reading this, I am not quite sure... Should the standard work with test kits? I just got the monitor, but I would love to see how my kits compare against the standards, if that will work. So far, against the tank water, my Lamotte kit is reading quite close to the monitor.
Thanks!
-Mike
Randy Holmes-Farley
01/30/2007, 10:40 AM
Yes, they should. :)
loudell
01/30/2007, 10:41 AM
The Nitrate standards should definately work with all colormetric nitrate test kits. Check the standards with your LaMotte Kit to see how 1 ppm, 10, ppm and 100 ppm will look relativel to color chart. If you are very precise on your measurement. You can combine 10 ppm nitrate standard with 1 ppm standard to create a standard that is 5.5 ppm NO3. I would suggest adding say 15 drops of each standard to a measurement vial...
mwitten
01/30/2007, 11:03 AM
Randy & Loudell,
Thanks! I'll give it a try tonight (at least the simple test).
-Mike
jamal-188
01/30/2007, 11:17 AM
I"m anxious to see how some of my test kits compare as well. Going to try an instant ocean, salifer, and aq pharm this week. I'll report back with the findings.
mwitten
02/01/2007, 07:38 PM
<a href=showthread.php?s=&postid=9113610#post9113610 target=_blank>Originally posted</a> by loudell
The Nitrate standards should definately work with all colormetric nitrate test kits. Check the standards with your LaMotte Kit to see how 1 ppm, 10, ppm and 100 ppm will look relativel to color chart. If you are very precise on your measurement. You can combine 10 ppm nitrate standard with 1 ppm standard to create a standard that is 5.5 ppm NO3. I would suggest adding say 15 drops of each standard to a measurement vial...
OK, well... Something ain't right, but I am not sure exactly what.
On the Lamotte kit, tried twice, I used the 10 ppm standard solution. The kit I have (low range) is SUPPOSED to give me a reading in Nitrate-N... thus I would have expected a reading of 10/4.4 or a bit over 2. It comes out well above 10... the highest range I can read on this kit.
I had ordered some additional solutions, so I did the second test with the it and had the same result (in case I had done something brilliant, like pouring the leftover 100 into the 10 bottle).
I also tried it with a HACH kit, but messed it up (was reading from the 0-1ppm Nitrate-N instructions, not 1-10), however with some basic dilution, it looked pretty close (very close, actually). I'll try both again this weekend, as well as the 1ppm solution to see how it goes. The Lamotte kit only has one set of instructions, so I don't think I could make the same mistake... especially twice.
I'll also verify my reagents, but I believe I replaced them just a few months ago.
If anyone sees a mistake I made, please let me know. I'd love to find out my Lamotte is reading way too high...javascript:smilie(':D')
-Mike
loudell
02/01/2007, 08:31 PM
Mike, FWIW you can not dilute the nitrate calibration matrix and use it with the nitrate meter. You may be able to use the diluted matrix with a color test kit, I can't say for sure, you may want to inquire with the manufacturer of the colormetric kit.
mwitten
02/02/2007, 05:43 AM
<a href=showthread.php?s=&postid=9136857#post9136857 target=_blank>Originally posted</a> by loudell
Mike, FWIW you can not dilute the nitrate calibration matrix and use it with the nitrate meter. You may be able to use the diluted matrix with a color test kit, I can't say for sure, you may want to inquire with the manufacturer of the colormetric kit.
Good to know for sure... and makes sense. I was going to ask, since the refills come in 3 bottle sets, and (with luck) I won't be doing much in the high range, making the 100ppm "spare".
That Hach kit has separate instructions for the 1-10 range, which I'll follow when I re-try: I totally messed that one up, and was just trying to salvage it to some degree. My only point was that the Hach kit had the potential to agree better than the LaMotte kit did and, as I understand it, the methods they use are similar. I did not dilute the sample for the LaMotte kit test.
Thanks!
-Mike
Randy Holmes-Farley
02/02/2007, 12:21 PM
The Hach kit should work diluted or not, as it is not intended only for salt water anyway, and the fact that it does work is good to hear, although not surprising. :)