I have a little dilemma in calibration. I bought a calibration solution at 35ppt. If I calibrate the refractometer using that solution and then check it against zero tds RO/DI water, they are off by 2ppt. Which is more accurate? Which should I use to calibrate?

allsps40
09/21/2010, 09:08 PM
Use the calibration fluid. Even Ro water is not 0ppt salinity. Found that out after I used Ro to calibrate my refractometer and SG was a bit high 1.028 when I thought it was 1.025.

goochesfish
09/21/2010, 09:36 PM
yes, that's what happens to me too. So i will use the calibration fluid. Thanks!

The Punisher
09/21/2010, 09:45 PM
I'd second going by the calibration fluid. The calibration fluid is a know variable that will give the best results.

jflick345
09/21/2010, 09:54 PM
Most refractometers actually come with instructions to NEVER calibrate them to 0. Use the solution, save yourself some headaches.

greenbean36191
09/22/2010, 07:35 AM
The issue isn't that your RO/DI water has a salinity of 2 ppt, but that the refractometer is off by 0.06 units per ppt. That inaccuracy adds up the further you get from the calibration point. If you calibrate it to 0 and then try to read at 32 ppt it will be off by almost 2 ppt. If you calibrate it to 35 ppt and try to read it at 32 ppt it will only be off by less than 0.2 ppt. A good rule of thumb is that if you're only using 1 calibration point, always choose the one closest to the values you want to measure.

sneeyatch
09/22/2010, 07:58 AM
A good rule of thumb is that if you're only using 1 calibration point, always choose the one closest to the values you want to measure.

Agreed - calibration solution.