I'm running a denitrator and trying to obtain negative orp values via my Apex. Unfortunately the Apex does not report values below 0 for orp. Based off the below post from an old thread on a similar issue. I believe I have come up with the formula to do this. Can anyone confirm? This makes me wish I paid attention in algebra.
7+(x/58)= y
X = orp value
Y= ph value using orp probe
As a work around, you should be able to use the ORP connector in pH mode. A pH reading above 7.0 is a negative ORP reading. For every 1 unit of pH change the ORP is changing by about 58 mV. An ORP reading of -200 is equivalent to a pH reading of about 7.0 + 200/58 = 10.44.
Curt
7+(x/58)= y
X = orp value
Y= ph value using orp probe
As a work around, you should be able to use the ORP connector in pH mode. A pH reading above 7.0 is a negative ORP reading. For every 1 unit of pH change the ORP is changing by about 58 mV. An ORP reading of -200 is equivalent to a pH reading of about 7.0 + 200/58 = 10.44.