Quote:
Originally Posted by slief
You hit the nail on the head.
His pump is a 10-600 RPM motor, not a 6-600 RPM. That head flows at 2.8 ml per revolution with LS24 tubing. At 10 RPM minimum on his pump, that is 28 ml/min minimum with LS24 tubing on that pump.
LS-15 tubing is 1.5 ml per revolution which would get him down to a minimum of 15 ml/min. Seems a change of tubing is in order if the target flow is below 28 ml/min.
|
Thank you for the breakdown and They have emailed me back and yes with ls24 it's 2.833 ml/min so that is why it cannot go below 30.
Sent from my SM-N950U using Tapatalk