Aragonite affecting pH probe
I am baffled.
I have a pH probe with my Apex. The probe is a Neptune lab grade, 3 months old. I added live sand ( aragalive?) a couple weeks ago and noticed my pH was 0.2 lower on average in the following days. After a week I recalibrated the probe and found the pH back to normal, the probe had been out of calibration. That was 2 days ago. Last night and moved some sand around. This created a little dust but did not badly cloud the water. The speed thing is my pH probe immediately flat lined and has not moved much in 11 hours, not filling my normal day/night trends. Why would the dust from the sand mess up the probe calibration? |
10/10/2017, 04:30 AM | #2 |
Registered Member
Join Date: Jan 2002
Location: Mississippi
Posts: 1,206
|
Not after a week...recalibrated after 2 weeks...
|
10/10/2017, 04:41 AM | #3 |
Registered Member
Join Date: Nov 2007
Location: Raleigh
Posts: 864
|
The little glss bulb on the tip of the probe works by H+ hydrating the thin membranes SiOH bonds only at the outter surface. The resistance across the glass bulb changes based on number of hydrated SiOHH+ but as you would guess that signal is very very weak to begin with. Cloud the probe surface up with who knows what from the sand and it will change the resistance/pH drastically. The glass bulb on a pH probe must be kept immaculately clean.
__________________
Acros, Zoas, Lps and Acros Kevin Current Tank Info: 200 gallon sps |
10/10/2017, 04:58 AM | #4 |
Registered Member
Join Date: Jan 2002
Location: Mississippi
Posts: 1,206
|
I had no idea it was that sensitive. Have not had sand in many years.
Do others remove their probes when doing water changes, etc? |
10/10/2017, 05:01 AM | #5 |
Registered Member
Join Date: Nov 2007
Location: Raleigh
Posts: 864
|
I don’t own a probe but I have monitored many a reaction with one. They foul very easy!
The electronics that boost the signal are very similar to an electric guitar amp Less distortion please
__________________
Acros, Zoas, Lps and Acros Kevin Current Tank Info: 200 gallon sps |
10/10/2017, 03:29 PM | #6 |
RC Mod
Join Date: Mar 2002
Location: Mountain View, CA, USA
Posts: 88,616
|
Water changes might cause a problem if there's a lot of debris stirred up, but I suspect that's unlikely. Adding sand is very likely to cause debris buildup, so I'm not that surprised about the problem. I would clean and calibrate the probe every month, and also I'd clean and calibrate it before taking any drastic action based on the measured pH level.
__________________
Jonathan Bertoni |
10/12/2017, 02:07 AM | #7 | |
Moved On
Join Date: Aug 2014
Location: Galveston Co.
Posts: 853
|
Quote:
Sent from my SAMSUNG-SM-G870A using Tapatalk |
|
10/12/2017, 05:05 AM | #8 |
Registered Member
Join Date: Nov 2007
Location: Raleigh
Posts: 864
|
I agree with you! pH probes are very innacurate/ not to be trusted without cleaning, they are just to see “trends”. I couldn’t have restated this thread better myself :3 thanks
The initial question was can my pH probe tip foul and get less accurate upon sand addition. What is your answer to that, yes or no?
__________________
Acros, Zoas, Lps and Acros Kevin Current Tank Info: 200 gallon sps Last edited by orcafood; 10/12/2017 at 05:11 AM. |
10/12/2017, 05:18 AM | #9 | |
Registered Member
Join Date: Jan 2002
Location: Mississippi
Posts: 1,206
|
Quote:
It's not the absolute number, but the fact that the probe is NOT trending as it should that raised my question. Thanks all for info. |
|
10/13/2017, 10:18 AM | #10 |
Moved On
Join Date: Aug 2014
Location: Galveston Co.
Posts: 853
|
Not trying to be smart was, my apologies..
Sent from my SM-G892A using Tapatalk |
10/13/2017, 10:24 AM | #11 |
Grizzled & Cynical
Join Date: Apr 2006
Location: Stamford, CT
Posts: 17,319
|
Trouble with any probe (maybe temp excepted) is that they are easily skewed by environmental crap (literally and figuratively). Dust, bubbles, electronic fields, etc. I experimented with different locations for mine (pH, Cond, ORP); ultimately concluding that the RDSB tank was the least 'disruptive'. You always want to be careful about what you control with probe readings.
__________________
Simon Got back into the hobby ..... planned to keep it simple ..... yeah, right ..... clearly I need a new plan! Pet peeve: anemones host clowns; clowns do not host anemones! Current Tank Info: 450 Reef; 120 refugium; 60 Frag Tank, 30 Introduction tank; multiple QTs |
10/14/2017, 02:16 AM | #12 |
Moved On
Join Date: Aug 2014
Location: Galveston Co.
Posts: 853
|
The probe , the double junction, imo only degraded in small , I mean small, increments, then it showed very unusual readings , clean in vinegar and it was close to original, recalibrated and ok, then it after that it was a a few more times then I replaced it, there good probes, I did have a New one that was bad, brs was replaced , my answer is no , I don't believe sediments effects ph probe that way, I can look at mine and see stuff on it, it's not newish looking it has algae on it. Just saying, calibration seems to be your problem.
Sent from my SM-G892A using Tapatalk |
10/14/2017, 02:17 AM | #13 |
Moved On
Join Date: Aug 2014
Location: Galveston Co.
Posts: 853
|
What kind of ph probe are you using
Sent from my SM-G892A using Tapatalk |
10/14/2017, 07:53 AM | #14 |
Registered Member
Join Date: Nov 2007
Location: Raleigh
Posts: 864
|
http://www.getbluelab.com/site/bluel...lean-DEC13.pdf
Be careful how you clean it The nicer the OP amp you have, the more you will notice a dirty probe. Probes themselves can be made dirt cheap, they are just a half cell with silver/ silver chloride. We could make one from an xmas bulb if we wanted haha
__________________
Acros, Zoas, Lps and Acros Kevin Current Tank Info: 200 gallon sps |
10/14/2017, 12:33 PM | #15 |
Registered Member
Join Date: Nov 2007
Location: Raleigh
Posts: 864
|
__________________
Acros, Zoas, Lps and Acros Kevin Current Tank Info: 200 gallon sps |
10/14/2017, 02:43 PM | #16 |
Moved On
Join Date: Aug 2014
Location: Galveston Co.
Posts: 853
|
Thought about it some more , it wasn't the dust , it may have been from the gas released from the sand bed
Sent from my SM-G892A using Tapatalk |
10/14/2017, 02:49 PM | #17 |
Registered Member
Join Date: Jan 2002
Location: Mississippi
Posts: 1,206
|
|
10/14/2017, 03:15 PM | #18 |
Moved On
Join Date: Aug 2014
Location: Galveston Co.
Posts: 853
|
Was the sand rinsed, might just be like dumping some kalk in the fine dust broke down.
Sent from my SM-G892A using Tapatalk |
10/15/2017, 05:22 AM | #19 |
Registered Member
Join Date: Jan 2002
Location: Mississippi
Posts: 1,206
|
Not rinsed, live sand in bag. It was in there two weeks and when I stirred it up, more dust came up and pH probe went wacky again.
|
10/15/2017, 08:27 AM | #20 |
Registered Member
Join Date: Aug 2008
Location: Cedar Park, Tx
Posts: 1,072
|
You do know CO2 has a dramatic effect on pH. You probably had your CO2 content in the water rise after adding the sand. When I have all 5 grandkids at the house my pH drops dramatically until they leave. I can usually tell when they come and go during the day. That drives my wife crazy when I tell her when they went out and come back during the day. It also went up considerably when we were gone for a few days. One person staying home will lower the pH for that day a little.
Sent from my iPad using Tapatalk |
10/15/2017, 08:30 AM | #21 |
Registered Member
Join Date: Mar 2013
Location: Central NC
Posts: 5,062
|
Keep in mind that there's always a possibility of "correlation without causation". pH probes will sometimes develop electrical connection problems in the cord, the electrode itself, or the BNC fitting. Moving the probe or otherwise messing with it can interrupt the connection between the meter and the probe, resulting in a "flat line" that doesn't respond to obvious changes in pH.
From the standpoint of the probe itself, I'd offer the following observations. Replacing the probe is best - they're cheap. If you'd rather not replace the probe and want to try cleaning it, I'd recommend the following: Remove the probe from the tank and rinse it. Contrary to the admonishment in the link Orca posted, rinsing with RODI will not damage the type of gel electrodes we use in our tank in any way, but tap water's OK as well. Prepare a solution of 0.1N hydrochloric acid - that's about a 1:100 dilution of concentrated hydrochloric acid sold for pool pH adjustment. Soak the bottom part of the probe in this acid solution for 30 minutes or so. Rinse thoroughly with tap or RODI, and recalibrate it with 7.0 and 10.0 pH buffer standards. The reason for using 0.1N HCl in this case is that it is extremely effective at removing calcium carbonate deposits, and will also denature and remove any protein coating on the pH electrode's junction. After performing the cleaning step, if the pH probe will not calibrate correctly, then it's trashed and must be replaced. |